Oh Dirk!
Thank you very much!!! There is no such thing as asynchronous messages with Tek scopes (the TPS-TDS series at least). Your second option worked out perfectly well! I must learn more about forward links and the db intricacies, though. But it's worth that. Thanks for help!!!
On Fri, Apr 27, 2012 at 11:39 AM, Dirk Zimoch <dirk.zimoch@psi.ch> wrote:
Hi Pavel,
This is indeed a bit tricky. You probably want the readback go to 1 when the scope is armed and go to 0 when the scope is done? Does the scope provide any asynchronous way to tell you when it is done? Unfortunately, I don't have a manual for this device. With GPIB, devices often can be programmed to send a SRQ when they are done with a time consuming activity. In that case the recipe is as follows:
Before arming the scope, set a status record to 1. When you get the "done" signal, set the status record to 0. You have to set the status before arming the scope so that you don't have a race condition if the scope is done very quickly. Thus you need an additional layer of records, one for the "user interface" and one for the "device interface" and set the status in between.
record(seq, "$(P)$(R)Arm") {
field(DESC, "Process this to start")
field(DOL1, "1")
field(LNK1, "$(P)$(R)ArmStatus PP")
field(DOL2, "1")
field(LNK2, "$(P)$(R)TrigArm PP")
record(bi, "$(P)$(R)ArmStatus") {
field(ZNAM, "STOP")
field(ONAM, "ARMED")
record(bo, "$(P)$(R)TrigArm") {
field(DESC, "Arm the trigger")
field(DTYP, "stream")
field(OUT, "@devTPS20xx.proto TrigArm $(PORT) $(A)")
field(ZNAM, "STOP")
record(bi, "$(P)$(R)WaitDone") {
field(INP, "@devTPS20xx.proto WaitDone $(PORT) $(A)")
field(SCAN, "I/O Intr")
field(FLNK, "$(P)$(R)SetDone")
record(seq, "$(P)$(R)SetDone") {
field(DOL1, "0")
TrigArm {
out "ACQ:STOPA SEQ";
out "ACQ:STATE %{OFF|ON}";
done=SRQ_code_for_done;
done_timeout=max_msec_to_wait_for_acquisition;
WaitDone {
event($done) $done_timeout;
out "ACQ:STATE?";
in "1";
But I have to admit that I have never really used the event command, because I don't have such a device. So it may not work :-)
If your device does not support asynchronous events, you probably have to poll. In this case the possible race condition is different and you have to arm the scope first, then start polling like this:
field(DOL1, "1")
field(LNK1, "$(P)$(R)TrigArm PP")
field(DOL2, "1")
field(LNK2, "$(P)$(R)ArmStatus PP")
field(ZNAM, "STOP")
field(ZNAM, "STOP")
record(bi, "$(P)$(R)PollDone") {
field(INP, "@devTPS20xx.proto PollDone $(PORT) $(A)")
field(SCAN, "1 second")
field(SDIS, "$(P)$(R)ArmStatus")
field(DISV, "0")
field(FLNK, "$(P)$(R)CheckDone")
record(calcout, "$(P)$(R)CheckDone"){
field(INPA, "$(P)$(R)PollDone")
field(CALC, "A")
field(OOPT, "When Zero")
field(OUT, "$(P)$(R)ArmStatus PP")
PollDone {
out "ACQ:STATE?";
in "%d";
Note that the PollDone polls only while ArmStatus is 1.
The poll rate depends on your application, of course.
Does this help?
Dirk
Pavel Masloff wrote:
Dear all,
I have a Tektronix scope (TPS2024). I am using StreamDevice. I know how to set the scope's parameters. These are simple commands with the SCAN field = PASSIVE.
One record (the Trigger Single Sequence button) is causing me some pain. When the scope is triggered (end of measurement), I need a readback signal to change its state (=0).
So far my commands were only one way (except for the initializaion). I send, nothing comes. Here is what I use:
field(SCAN, "Passive")
field(ZNAM, "STOP")
field(PINI, "")
TrigArm {
ReplyTimeout = 2000;
out "ACQ:STATE %{OFF|ON}"; @init {
out "ACQ:STATE?";
in "%i";
So I need the same @init code in the TrigArm protocol (out "ACQ:STATE?";in "%i";). How can I do that?
The "ACQ:STATE?" command will have an answer ("1"=Trigger Armed, "0"=Trigger Not Armed).
I have tried using another record (with forward link), but to no avail.
-- Best regards,
Pavel Maslov, MS Controls Engineer at Pulsed power Lab Efremov Institute for Electro-Physical Apparatus
St. Petersburg, Russia
Mobile: +7 (951) 672 22 19
Landline: +7 (812) 461 01 01
- [StreamDevice] need advice Pavel Masloff
ECHO variable in build rules Benjamin Franksen
[Scopes] BMP image record?? Pavel Masloff
<2012>
Problem installing EDM on Fedora 16 manta santosh aditya santosh_aditya
<2012>
