1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  <20172018  2019  2020  2021  2022  2023  2024  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  <20172018  2019  2020  2021  2022  2023  2024
Subject: RE: new Scaler driver
To: "'[email protected]'" <[email protected]>, tech-talk <[email protected]>
Date: Thu, 21 Sep 2017 18:44:24 +0000
I think your driver is close to what you want, but there are a few problems.  These have to do with the set timer command.
First, you are sending the comment TPRdddd where dddd is the time in ms.  However, according to their manual this is incorrect.  You send the command TPR? to read back the timer value, but to set the timer value you have to send the command STPRdddd, i.e. the command is STPR, not TPR.
Also, you are just sending 0 for the value of the timer.  You need set a value for the clock frequency of your device.  Do you know what this is?  If not, the actual value is not critical, you could just say that it is 10MHz (10000000).  Both your driver and the scaler record need to agree on the value for the frequency, i.e. FREQUENCY below must agree with the .FREQ field in the scaler record.
To set the preset time you then take the preset counts that are passed in writeInt32 and divide by the frequency and multiple by 1000 to convert from seconds to ms.
So your code would change to:
#define FREQUENCY 10000000.
    else if(function==this->scalerPreset)   //  modified a lot
            int m;  // m means millsecond
            char newstr[40];
            m =  1000. * value / FREQUENCY
            sprintf(newstr,"STPR%d",m);
            this->sendCommand(newstr, response, sizeof(response), &responseLen);
            asynPrint(pasynUser, ASYN_TRACE_FLOW, "%s;%s scalerPreset channel", driverName, functionName);
From: [email protected] [mailto:[email protected]] On Behalf Of ???
Sent: Thursday, September 21, 2017 9:40 AM
To: tech-talk
Subject: new Scaler driver
Hi all,
  I am trying to modified drvScaler974.cpp( driver of 974 scaler ) to drvScalerCT08E.cpp in std of synapps.
CT08E was a new scaler( http://www.tsuji-denshi.co.jp/english/download.html ), which has not driver in std.
  In a word, I want to write the driver of CT08E base on drvScaler974.cpp in std-3-4.
  The command was uncomplex. I modified the relevant command, such as clear,start and stop. But command of new scaler about "set timer", "get value of channel" was much different from 974scaler.
The new scaler command about "set timer": If I sent TPR%d(%d was millsecond), the scaler would set the timer.
The new scaler command about "get value of channel": If I sent RDAL, I should get 9 number(channel 0-7 and the timer), whose format was 0000000000 000000000 0000000000 0000000000 0000000000 0000000000 0000000000 0000000000 0000000000, nine 10 bit number divide by space. And the laset number(timer) continue increase until it equal the set time which TPR command set.
Rebuild std and ioc was ok, the ioc could run with no error except the write/read.
I am using base 3.15.2, Synapps5_8, std-3-4, Centos 7(64).
The driver file and output was following.
How could I do it to achieve these function?
Could you give me some suggestion.
Thank you,
Liang yx
The content of drvScalerCT08E.cpp which I modified.
#include <stdlib.h>
#include <stddef.h>
#include <string.h>
#include <stdio.h>
#include <math.h>
/* EPICS includes */
#include <asynPortDriver.h>
#include <asynOctetSyncIO.h>
#include <epicsEvent.h>
#include <epicsThread.h>
#include <epicsTime.h>
#include <epicsExport.h>
#include <errlog.h>
#include <iocsh.h>
#include "devScalerAsyn.h"
#define MAX_CHANNELS 8                    //////////////////////////chang 4 to 8
#define timeOut 0.1
static const char *driverName= "ScalerCT08E";  // >- Scaler974 change to CT08E
class ScalerCT08E:public asynPortDriver
public:
    ScalerCT08E(const char *portName, const char *serialPort, int serialAddr, int poll);
    virtual asynStatus writeInt32(asynUser *pasynUser, epicsInt32 value);
    virtual asynStatus readInt32Array(asynUser *pasynUser,epicsInt32 *value, size_t nElements, size_t *nIn);
    virtual void report(FILE *fp, int details);
    void eventThread();
private:
    int scalerReset;
    int scalerChannels;
    int scalerRead;
    int scalerReadSingle;
    int scalerPreset;
    int scalerArm;
    int scalerDone;
    double polltime;
    epicsEventId eventId;
    asynUser *pasynUserScaler;
    asynStatus sendCommand(const char *command, char *statusString, size_t maxStatusLen, size_t *statusLen);
    asynStatus sendCommand(const char *command, char *statusString, size_t maxStatusLen, size_t *statusLen,
                           char *response, size_t maxResponseLen, size_t *responseLen);
static void eventThreadC(void *pPvt)
    ScalerCT08E *pScalerCT08E = (ScalerCT08E *)pPvt;
    pScalerCT08E->eventThread();
ScalerCT08E::ScalerCT08E(const char *portName, const char *serialPort, int serialAddr, int poll)
    :asynPortDriver(portName, MAX_CHANNELS, 7,
                    asynInt32Mask | asynInt32ArrayMask | asynDrvUserMask,
                    asynInt32Mask,
                    /* Should also be ASYN_CANBLOCK, but device support does not work with asynchronous devices */
                    ASYN_MULTIDEVICE,1,0,0)
    int i;
    asynStatus status;
    static const char *functionName="ScalerCT08E";
    if (poll==0) poll=100;
    this->polltime=poll / 1000.;
    this->eventId = epicsEventCreate(epicsEventEmpty);
    status = pasynOctetSyncIO->connect(serialPort, serialAddr, &this->pasynUserScaler, NULL);
    if (status != asynSuccess) {
        asynPrint(this->pasynUserSelf, ASYN_TRACE_ERROR,
            "%s:%s: error connecting to port %s address %d\n",
            driverName, functionName, serialPort, serialAddr);
        return;
    createParam(SCALER_RESET_COMMAND_STRING,        asynParamInt32,     &this->scalerReset);
    createParam(SCALER_CHANNELS_COMMAND_STRING,     asynParamInt32,     &this->scalerChannels);
    createParam(SCALER_READ_COMMAND_STRING,         asynParamInt32Array,&this->scalerRead);
    createParam(SCALER_READ_SINGLE_COMMAND_STRING,  asynParamInt32,     &this->scalerReadSingle);
    createParam(SCALER_PRESET_COMMAND_STRING,       asynParamInt32,     &this->scalerPreset);
    createParam(SCALER_ARM_COMMAND_STRING,          asynParamInt32,     &this->scalerArm);
    createParam(SCALER_DONE_COMMAND_STRING,         asynParamInt32,     &this->scalerDone);
    setIntegerParam(scalerChannels, MAX_CHANNELS);
    setIntegerParam(scalerDone, 1);
    for (i=0; i<MAX_CHANNELS; i++) setIntegerParam(i, scalerReadSingle, 0);
    epicsThreadCreate("ScalerCT08E",
                      epicsThreadPriorityMedium,
                      epicsThreadGetStackSize(epicsThreadStackMedium),
                      (EPICSTHREADFUNC)eventThreadC,this);
asynStatus ScalerCT08E::sendCommand(const char *command, char *statusString, size_t maxStatusLen, size_t *statusLen)
    return sendCommand(command, statusString, maxStatusLen, statusLen, NULL, 0, NULL);
asynStatus ScalerCT08E::sendCommand(const char *command, char *statusString, size_t maxStatusLen, size_t *statusLen,
    size_t nWrite;
    asynStatus status;
    double timeout = 1.0;
    int eomReason;
    static const char *functionName = "sendCommand";
    if (response != NULL) {
        status = pasynOctetSyncIO->writeRead(this->pasynUserScaler, command, strlen(command),
                                             response, maxResponseLen, timeout, &nWrite, responseLen, &eomReason);
        if (status != asynSuccess) goto done;
        status = pasynOctetSyncIO->read(this->pasynUserScaler,
                                        statusString, maxStatusLen, timeout, statusLen, &eomReason);
    } else {
                                             statusString, maxStatusLen, timeout, &nWrite, statusLen, &eomReason);
    done:
                  "%s:%s: writing command %s, error=%s\n",
                  driverName, functionName, command, this->pasynUserScaler->errorMessage);
    return(status);
asynStatus ScalerCT08E::writeInt32(asynUser *pasynUser, epicsInt32 value)
    int function = pasynUser->reason;
    char response[512];   //increase the number 256 ->512
    size_t responseLen;
    asynStatus status = asynSuccess;
    static const char *functionName="writeInt32";
    setIntegerParam(function, value);
    if(function==this->scalerReset)
        this->sendCommand("STOP", response, sizeof(response), &responseLen);
        this->sendCommand("CLAL", response, sizeof(response), &responseLen);    //  i think here is the command, so I  changed  CLEAR_ALL to CLAL
        asynPrint(pasynUser, ASYN_TRACE_FLOW, "%s:%s scalerReset\n",driverName, functionName);
    else if(function==this->scalerArm)
        if(value !=0)
        /* Start counting */
            this->sendCommand("STRT", response, sizeof(response), &responseLen);  // START TO STRT
            setIntegerParam(scalerDone, 0);
            epicsEventSignal(this->eventId);
        else
        /* Stop counting */
            status = this->sendCommand("STOP", response, sizeof(response), &responseLen);
        asynPrint(pasynUser, ASYN_TRACE_FLOW, "%s:%s scalerArm=%d\n", driverName, functionName,value);
            char newstr[40];
            sprintf(newstr,"TPR%d",m);
//            setIntegerParam(scalerDone, 0);
        int m,n;
        char newstr[20];
        n=(int)log10(double(value));
        m=(int)(value/pow(10.0,n));
        sprintf(newstr, "SET_COUNT_PRESET %d,%d", m, n);
    else
        asynPrint(pasynUser,ASYN_TRACE_ERROR,"%s:%s got illegal function %d\n", driverName, functionName, function);
    callParamCallbacks();
    return(status);
asynStatus ScalerCT08E::readInt32Array(asynUser *pasynUser, epicsInt32 *value, size_t maxChannel, size_t *nIn)
    static const char *functionName="readInt32Array";
    int temp;
    int i;
    if (maxChannel > MAX_CHANNELS) maxChannel = MAX_CHANNELS;
    *nIn = maxChannel;
    if (function==scalerRead)
        for (i=0; i<(int)maxChannel; i++)
            getIntegerParam(i, scalerReadSingle, &temp);
            value[i] = temp;
        asynPrint(pasynUser, ASYN_TRACEIO_DRIVER,
            "%s:%s: value=%d %d %d %d\n",
            driverName, functionName, value[0], value[1], value[2], value[3]);
    } else {
        return(asynError);
    return(asynSuccess);
void ScalerCT08E::report(FILE *fp, int details)
    asynPortDriver::report(fp, details);
void ScalerCT08E::eventThread()
    int done, presetCount;
    char response[100], statusString[20];
    size_t responseLen, statusLen;
    int counts[MAX_CHANNELS];
    int i;
    asynStatus status;
    static const char *functionName="eventThread";
    while(1)      //  SHOW_COUNTS need to change a lot,later
        epicsEventWait(this->eventId);
        while(1)
            status = this->sendCommand("RDAL", statusString, sizeof(statusString), &statusLen,
            sscanf(response, "%d %d %d %d %d %d %d %d %d",
                   &counts[0], &counts[1], &counts[2], &counts[3], &counts[4], &counts[5], &counts[6], &counts[7], &counts[8]);  //counts[8] was the timer
            asynPrint(this->pasynUserSelf, ASYN_TRACEIO_DRIVER,
                "%s:%s status=%d, counts=%d %d %d %d %d %d %d %d %d\n",
                driverName, functionName, status, counts[0], counts[1], counts[2], counts[3], counts[4], counts[5], counts[6], counts[7], counts[8] );     // I think here should be changed, I added "%d".
            this->lock();
            /* Get value of done in case scaler was stopped by scalerArm(0) */
            getIntegerParam(scalerDone, &done);
            getIntegerParam(scalerPreset, &presetCount);
            if (!done && (counts[0] >= presetCount)) done = 1;
            setIntegerParam(scalerDone, done);
            for (i=0; i<MAX_CHANNELS; i++) {
                setIntegerParam(i, scalerReadSingle, counts[i]);
                callParamCallbacks(i, i);
            this->unlock();
            if (done) break;
            epicsThreadSleep(this->polltime);
extern "C" int initScalerCT08E(const char *portName, const char *serialPort, int serialAddr, int poll)
    new ScalerCT08E(portName, serialPort, serialAddr, poll);
/* iocsh function */
static const iocshArg initArg0 = {"Port Name", iocshArgString};
static const iocshArg initArg1 = {"Scaler Port", iocshArgString};
static const iocshArg initArg2 = {"GPIB address", iocshArgString};
static const iocshArg initArg3 = {"Poll", iocshArgInt};
static const iocshArg *const initArgs[] = {&initArg0,
                                           &initArg1,
                                           &initArg2,
                                           &initArg3
static const iocshFuncDef initFuncDef = {"initScalerCT08E",4, initArgs}; // 4 means 4 channel? I have 8 channel, maybe not ,just the number of args.
static void initCallFunc(const iocshArgBuf *args)
    initScalerCT08E(args[0].sval, args[1].sval, args[2].ival, args[3].ival);  //init command
void ScalerCT08ERegister(void)
    iocshRegister(&initFuncDef,initCallFunc);
extern "C" {
epicsExportRegistrar(ScalerCT08ERegister);
The information of start IOC :
Successfully locked memory using mlockAll
# Linux startup script
# For devIocStats
#epicsEnvSet("ENGINEER","engineer")
#epicsEnvSet("LOCATION","location")
#epicsEnvSet("GROUP","group")
< envPaths
epicsEnvSet("ARCH","linux-x86_64")
epicsEnvSet("IOC","iocLinux")
epicsEnvSet("TOP","/opt/synApps_5_8/support/xxx-5-8-3")
epicsEnvSet("SUPPORT","/opt/synApps_5_8/support")
epicsEnvSet("EPICS_BASE","/opt/base")
epicsEnvSet("ALIVE","/opt/synApps_5_8/support/alive-1-0")
epicsEnvSet("ALLEN_BRADLEY","/opt/synApps_5_8/support/allenBradley-2-3")
epicsEnvSet("AREA_DETECTOR","/opt/synApps_5_8/support/areaDetector-R2-0")
epicsEnvSet("ADCORE","/opt/synApps_5_8/support/areaDetector-R2-0/ADCore-R2-2")
epicsEnvSet("ADBINARIES","/opt/synApps_5_8/support/areaDetector-R2-0/ADBinaries-R2-2")
epicsEnvSet("ASYN","/opt/synApps_5_8/support/asyn-4-32")
epicsEnvSet("AUTOSAVE","/opt/synApps_5_8/support/autosave-5-6-1")
epicsEnvSet("BUSY","/opt/synApps_5_8/support/busy-1-6-1")
epicsEnvSet("CALC","/opt/synApps_5_8/support/calc-3-4-2-1")
epicsEnvSet("CAMAC","/opt/synApps_5_8/support/camac-2-7")
epicsEnvSet("CAPUTRECORDER","/opt/synApps_5_8/support/caputRecorder-1-6")
epicsEnvSet("DAC128V","/opt/synApps_5_8/support/dac128V-2-8")
epicsEnvSet("DELAYGEN","/opt/synApps_5_8/support/delaygen-1-1-1")
epicsEnvSet("DEVIOCSTATS","/opt/synApps_5_8/support/devIocStats-3-1-13")
epicsEnvSet("DXP","/opt/synApps_5_8/support/dxp-3-4")
epicsEnvSet("IP","/opt/synApps_5_8/support/ip-2-17")
epicsEnvSet("IP330","/opt/synApps_5_8/support/ip330-2-8")
epicsEnvSet("IPAC","/opt/synApps_5_8/support/ipac-2-13")
epicsEnvSet("IPUNIDIG","/opt/synApps_5_8/support/ipUnidig-2-10")
epicsEnvSet("LOVE","/opt/synApps_5_8/support/love-3-2-5")
epicsEnvSet("MCA","/opt/synApps_5_8/support/mca-7-6")
epicsEnvSet("MEASCOMP","/opt/synApps_5_8/support/measComp-1-1")
epicsEnvSet("MODBUS","/opt/synApps_5_8/support/modbus-2-7")
epicsEnvSet("MOTOR","/opt/synApps_5_8/support/motor-6-9")
epicsEnvSet("OPTICS","/opt/synApps_5_8/support/optics-2-9-3")
epicsEnvSet("QUADEM","/opt/synApps_5_8/support/quadEM-5-0")
epicsEnvSet("SNCSEQ","/opt/synApps_5_8/support/seq-2-2-3")
epicsEnvSet("SOFTGLUE","/opt/synApps_5_8/support/softGlue-2-4-3")
epicsEnvSet("SSCAN","/opt/synApps_5_8/support/sscan-2-10-1")
epicsEnvSet("STD","/opt/synApps_5_8/support/std-3-4")
epicsEnvSet("STREAM","/opt/synApps_5_8/support/stream-2-6a")
epicsEnvSet("VAC","/opt/synApps_5_8/support/vac-1-5-1")
epicsEnvSet("VME","/opt/synApps_5_8/support/vme-2-8-2")
epicsEnvSet("XXX","/opt/synApps_5_8/support/xxx-5-8-3")
epicsEnvSet(("STREAM_PROTOCOL_PATH","/opt/synApps_5_8/support/xxx-5-8-3")
# save_restore.cmd needs the full path to the startup directory, which
# envPaths currently does not provide
epicsEnvSet(STARTUP,/opt/synApps_5_8/support/xxx-5-8-3/iocBoot/iocLinux)
# Increase size of buffer for error logging from default 1256
errlogInit(20000)
# Specify largest array CA will transport
# Note for N doubles, need N*8 bytes+some overhead
epicsEnvSet EPICS_CA_MAX_ARRAY_BYTES 64010
# Tell EPICS all about the record types, device-support modules, drivers,
# etc. in the software we just loaded (xxx.munch)
dbLoadDatabase("../../dbd/iocxxxLinux.dbd")
iocxxxLinux_registerRecordDeviceDriver(pdbbase)
### save_restore setup
#< save_restore.cmd
# Access Security
dbLoadRecords("/opt/synApps_5_8/support/xxx-5-8-3/xxxApp/Db/Security_Control.db","P=xxx:")
asSetFilename("/opt/synApps_5_8/support/xxx-5-8-3/iocBoot/accessSecurity.acf")
asSetSubstitutions("P=xxx:")
# trap listener
#dbLoadRecords("$(CAPUTRECORDER)/caputRecorderApp/Db/caputPoster.db","P=xxx:,N=300")
#doAfterIocInit("registerCaputRecorderTrapListener('xxx:caputRecorderCommand')")
# GUI database
#dbLoadRecords("$(CAPUTRECORDER)/caputRecorderApp/Db/caputRecorder.db","P=xxx:,N=300")
# soft scaler for testing
#< softScaler.cmd
# user-assignable ramp/tweak
###dbLoadRecords("$(STD)/stdApp/Db/ramp_tweak.db","P=xxx:,Q=rt1")
drvAsynIPPortConfigure("CT08E", "192.168.1.123:7777", 0, 0, 0)
asynOctetSetInputEos("CT08E",0,"\r\n")
asynOctetSetOutputEos("CT08E",0,"\r\n")
asynSetTraceIOMask("CTStarting iocInit
08E",0,0x2)
asynSetTraceMask("CT08E",0,255)
asynOctetConnect("CT08E","CT08E")
#asynOctetWriteRead("CT08E","INIT")
#asynSetTraceIOMask("CT08E",-1,0x9)
initScalerCT08E("serialCT08E","CT08E",0,100)   ####
### Stuff for user programming ###
###< calc.cmd
# 4-step measurement
#dbLoadRecords("$(STD)/stdApp/Db/4step.db", "P=xxx:,Q=4step:")
# Miscellaneous PV's, such as burtResult
dbLoadRecords("/opt/synApps_5_8/support/std-3-4/stdApp/Db/misc.db","P=xxx:")
dbLoadRecords("/opt/synApps_5_8/support/std-3-4/stdApp/Db/scaler.db","P=xxx:,S=scaler1,OUT=@asyn(serialCT08E 0 0),DTYP=Asyn Scaler,FREQ=10,TP=1,TP1=0.5,PR1=1,CONT=1")
iocInit
## EPICS R3.15.2 $Date: Thu 2015-05-14 14:09:28 +0200$
## EPICS Base built Oct 16 2016
# Report  states of database CA links
dbcar(*,1)
CA links in all records
Total 0 CA links; 0 connected, 0 not connected.
    0 can't read, 0 can't write.  (0 disconnects, 0 writes prohibited)
I used scaler.opi.
After I pressed the "Count" button, the output of IOC:
epics> 2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort created queueLockPortPvt=0x7f6a2c047880
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort created queueLockPortPvt=0x7f6a2c047880, event=0x7f6a2c0478a0, mutex=0x7f6a2c047950
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort taking mutex 0x7f6a2c047950
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:18.298 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:18.298 asynManager::portThread port=CT08E callback
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:18.298 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:18.298 CT08E flush
2017/09/21 17:40:18.298 192.168.1.123:7777 flush
2017/09/21 17:40:18.298 192.168.1.123:7777 write.
2017/09/21 17:40:18.298 192.168.1.123:7777 write 6
STOP\r\n
2017/09/21 17:40:18.298 wrote 6 to 192.168.1.123:7777, return asynSuccess.
2017/09/21 17:40:18.298 CT08E wrote
STOP\r\n
2017/09/21 17:40:18.298 asynOctetSyncIO wrote:
STOP
2017/09/21 17:40:18.298 192.168.1.123:7777 read.
2017/09/21 17:40:19.298 CT08E queueUnlockPort
2017/09/21 17:40:19.298 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:19.298 CT08E queueUnlockPort unlock mutex 0x7f6a2c047950 complete.
2017/09/21 17:40:19.298 ScalerCT08E:sendCommand: writing command STOP, error=192.168.1.123:7777 timeout: Resource temporarily unavailable
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPort taking mutex 0x7f6a2c047950
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:19.298 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:19.298 asynManager::portThread port=CT08E callback
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:19.298 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:19.298 CT08E flush
2017/09/21 17:40:19.298 192.168.1.123:7777 flush
2017/09/21 17:40:19.298 192.168.1.123:7777 write.
2017/09/21 17:40:19.298 192.168.1.123:7777 write 6
CLAL\r\n
2017/09/21 17:40:19.298 wrote 6 to 192.168.1.123:7777, return asynSuccess.
2017/09/21 17:40:19.298 CT08E wrote
CLAL\r\n
2017/09/21 17:40:19.298 asynOctetSyncIO wrote:
CLAL
2017/09/21 17:40:19.298 192.168.1.123:7777 read.
2017/09/21 17:40:20.299 CT08E queueUnlockPort
2017/09/21 17:40:20.299 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:20.299 CT08E queueUnlockPort unlock mutex 0x7f6a2c047950 complete.
2017/09/21 17:40:20.299 ScalerCT08E:sendCommand: writing command CLAL, error=192.168.1.123:7777 timeout: Resource temporarily unavailable
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPort taking mutex 0x7f6a2c047950
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:20.299 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:20.299 asynManager::portThread port=CT08E callback
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:20.299 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:20.299 CT08E flush
2017/09/21 17:40:20.299 192.168.1.123:7777 flush
2017/09/21 17:40:20.299 192.168.1.123:7777 write.
2017/09/21 17:40:20.299 192.168.1.123:7777 write 6
TPR0\r\n
2017/09/21 17:40:20.299 wrote 6 to 192.168.1.123:7777, return asynSuccess.
2017/09/21 17:40:20.299 CT08E wrote
TPR0\r\n
2017/09/21 17:40:20.299 asynOctetSyncIO wrote:
TPR0
2017/09/21 17:40:20.299 192.168.1.123:7777 read.
2017/09/21 17:40:21.299 CT08E queueUnlockPort
2017/09/21 17:40:21.299 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:21.299 CT08E queueUnlockPort unlock mutex 0x7f6a2c047950 complete.
2017/09/21 17:40:21.299 ScalerCT08E:sendCommand: writing command TPR0, error=192.168.1.123:7777 timeout: Resource temporarily unavailable
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPort taking mutex 0x7f6a2c047950
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:21.299 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:21.299 asynManager::portThread port=CT08E callback
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:21.299 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:21.299 CT08E flush
2017/09/21 17:40:21.299 192.168.1.123:7777 flush
2017/09/21 17:40:21.299 192.168.1.123:7777 write.
2017/09/21 17:40:21.299 192.168.1.123:7777 write 6
STRT\r\n
2017/09/21 17:40:21.299 wrote 6 to 192.168.1.123:7777, return asynSuccess.
2017/09/21 17:40:21.299 CT08E wrote
STRT\r\n
2017/09/21 17:40:21.299 asynOctetSyncIO wrote:
STRT
2017/09/21 17:40:21.299 192.168.1.123:7777 read.
2017/09/21 17:40:22.299 CT08E queueUnlockPort
2017/09/21 17:40:22.299 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:22.299 CT08E queueUnlockPort unlock mutex 0x7f6a2c047950 complete.
2017/09/21 17:40:22.299 ScalerCT08E:sendCommand: writing command STRT, error=192.168.1.123:7777 timeout: Resource temporarily unavailable
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort created queueLockPortPvt=0x7f6a300008c0
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort created queueLockPortPvt=0x7f6a300008c0, event=0x7f6a300008e0, mutex=0x7f6a30000990
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort taking mutex 0x7f6a30000990
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:22.299 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:22.299 asynManager::portThread port=CT08E callback
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:22.299 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:22.299 CT08E flush
2017/09/21 17:40:22.299 192.168.1.123:7777 flush
2017/09/21 17:40:22.299 192.168.1.123:7777 write.
2017/09/21 17:40:22.299 192.168.1.123:7777 write 6
RDAL\r\n
2017/09/21 17:40:22.300 wrote 6 to 192.168.1.123:7777, return asynSuccess.
2017/09/21 17:40:22.300 CT08E wrote
RDAL\r\n
2017/09/21 17:40:22.300 asynOctetSyncIO wrote:
RDAL
2017/09/21 17:40:22.300 192.168.1.123:7777 read.
2017/09/21 17:40:22.303 192.168.1.123:7777 read 100
0000000004 0000000000 0000000000 0000000000 0000000000 0000000000 0000000000 000
2017/09/21 17:40:22.303 CT08E read
2017/09/21 17:40:22.303 asynOctetSyncIO read:
2017/09/21 17:40:22.303 CT08E queueUnlockPort
2017/09/21 17:40:22.303 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:22.303 CT08E queueUnlockPort unlock mutex 0x7f6a30000990 complete.
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPort taking mutex 0x7f6a30000990
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:22.303 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:22.303 asynManager::portThread port=CT08E callback
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:22.303 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:22.303 192.168.1.123:7777 read.
2017/09/21 17:40:23.303 CT08E queueUnlockPort
2017/09/21 17:40:23.303 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:23.303 CT08E queueUnlockPort unlock mutex 0x7f6a30000990 complete.
2017/09/21 17:40:23.303 ScalerCT08E:sendCommand: writing command RDAL, error=192.168.1.123:7777 timeout: Resource temporarily unavailable
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPort taking mutex 0x7f6a30000990
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:23.403 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:23.403 asynManager::portThread port=CT08E callback
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
2017/09/21 17:40:23.403 CT08E asynManager::queueLockPort got event from callback
2017/09/21 17:40:23.403 CT08E flush
2017/09/21 17:40:23.403 192.168.1.123:7777 flush
2017/09/21 17:40:23.403 192.168.1.123:7777 write.
2017/09/21 17:40:23.403 192.168.1.123:7777 write 6
RDAL\r\n
2017/09/21 17:40:23.403 wrote 6 to 192.168.1.123:7777, return asynSuccess.
2017/09/21 17:40:23.403 CT08E wrote
RDAL\r\n
2017/09/21 17:40:23.403 asynOctetSyncIO wrote:
RDAL
2017/09/21 17:40:23.403 192.168.1.123:7777 read.
2017/09/21 17:40:23.407 192.168.1.123:7777 read 100
0000000007 0000000000 0000000000 0000000000 0000000000 0000000000 0000000000 000
2017/09/21 17:40:23.407 CT08E read
2017/09/21 17:40:23.407 asynOctetSyncIO read:
2017/09/21 17:40:23.407 CT08E queueUnlockPort
2017/09/21 17:40:23.407 CT08E asynManager::queueUnlockPort waiting for event
2017/09/21 17:40:23.407 CT08E queueUnlockPort unlock mutex 0x7f6a30000990 complete.
2017/09/21 17:40:23.407 CT08E asynManager::queueLockPort locking port
2017/09/21 17:40:23.407 CT08E asynManager::queueLockPort taking mutex 0x7f6a30000990
2017/09/21 17:40:23.407 CT08E asynManager::queueLockPort queueing request
2017/09/21 17:40:23.407 CT08E addr -1 queueRequest priority 0 not lockHolder
2017/09/21 17:40:23.407 CT08E asynManager::queueLockPort waiting for event
2017/09/21 17:40:23.407 asynManager::portThread port=CT08E callback
2017/09/21 17:40:23.407 CT08E asynManager::queueLockPortCallback signaling begin event
2017/09/21 17:40:23.407 CT08E asynManager::queueLockPortCallback waiting for mutex from queueUnlockPort
new Scaler driver 梁雅翔
Prev: Re: Enumerated Strings in CSS BOY Kasemir, Kay
Next: RE: Enumerated Strings in CSS BOY Iain Marcuson
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  <20172018  2019  2020  2021  2022  2023  2024
Prev: new Scaler driver 梁雅翔
Next: APS Firewall Dudley, David
ANJ, 21 Dec 2017 Valid HTML 4.01! · Home · News · About · Base · Modules · Extensions · Distributions · Download ·