Hi Rod,
Not specifying any terminator means: Do not touch the terminator
settings. That means use whatever is specified by asynDriver with
asynOctetSetInputEos and asynOctetSetOutputEos in the startup script. If
neither is used, the default is empty string for both. Setting the
terminator to "" or any other string overwrites the asynDriver setting.
I had a look at drvLinuxGpib.c and found:
* It never sets eomReason. Thus StreamDevice (or any other asynUser) has
no chance to find EOI.
* Under some circumstances it sets EOS to '\0' instead of nothing.
Dirk
Rod Nussbaumer wrote:
Thanks, Dirk.
I have observed the 64-byte reads often in the past, and wondered
whether there was a way to increase that amount, for purposes of
efficiency. In this case, I do know the size of the array I expect to
read, as it is a fixed size, and each binary word is the same, plus some
accompanying header data, also of fixed length and makeup.
I am reading/matching the trailing LF. I wasn't sure whether setting
InTerminator = ""
is the same as not setting it explicitly at all, or whether it might
cause it to take on some new meaning. I do know that each protocol can
have specific terminator settings, and I've been trying to use that in
various combinations. What is the expected behavior if no Terminator or
InTerminator is specified, neither globally within the protocol file,
nor in the specific protocol? What I'm seeing in that case appears that
any NULL byte terminates a read. This causes a huge number of short
reads per waveform.
2012/01/13 08:50:10.847 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.847 GPIB0 addr 10 gpibPortRead
24 00
2012/01/13 08:50:10.857 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.857 GPIB0 addr 10 gpibPortRead
00
2012/01/13 08:50:10.868 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.868 GPIB0 addr 10 gpibPortRead
00
2012/01/13 08:50:10.878 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.878 GPIB0 addr 10 gpibPortRead
P
50 00
2012/01/13 08:50:10.888 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.889 GPIB0 addr 10 gpibPortRead
00
2012/01/13 08:50:10.899 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.899 GPIB0 addr 10 gpibPortRead
0a
2012/01/13 08:50:10.909 GPIB0 addr 10 gpibRead
2012/01/13 08:50:10.929 GPIB0 addr 10 : device timed out.
epics>
If I set InTerminator="" in the protocol, the same thing happens, but I
also get, at the end of the record processing:
00
2012/01/13 08:54:42.269 GPIB0 addr 10 gpibRead
2012/01/13 08:54:42.269 GPIB0 addr 10 gpibPortRead
0a
2012/01/13 08:54:42.279 GPIB0 addr 10 gpibRead
2012/01/13 08:54:42.299 GPIB0 addr 10 : device timed out.
2012/01/13 08:54:42.299 GPIB0 addr 10 setEos eoslen 0
Note the reference to EOS in the last asyn message. I sense that this is
germane to the problem.
I am a bit curious about *how* terminators are used/implemented in
streamDevice. Does the recognition of terminators get used only within
streamDevice, or does it get passed down to asyn in some way? To me,
there are different purposes to the terminators, especially in the case
of GPIB, where the low level drivers have their own behaviors for
message termination. It seems that there may be some interplay that is
hidden.
So far I think I've determined (from code inspections in the asyn linux
GPIB port driver) that the behavior of asyn plus the asynSetOptions()
I'm using should be mimicking my own test/diagnostic C code. I wonder
whether streamDevice is asserting some communication options through its
use of asyn? I will start looking into the streamDevice code, but if
anyone knows for sure, I would be happy to hear.
Mark, I haven't tried the asyn record tests yet, as I'm not really
familiar with how to implement and use it, but now is looking like a
good time learn.
Thanks.
--- rod.
Dirk Zimoch wrote:
Hi Rod,
If everything works correctly in asyn, it sees the EOI and tells
StreamDevice that this is the end of the message via the (eomReason &
ASYN_EOM_END) bit (in opposite to the ASYN_EOM_EOS bit when a terminator
is found). So it should work with an empty terminator.
I guess you know that you can set the terminators individually for each
protocol. So ASCII protocols can use LF and binary protocols "". In
order to get rid of the LF at the end of the message, you have to read
it in the protocol after reading the array. The waveform record reads
data until either NELM elements are read or the message ends. If the
array has variable size, you need to set NELM large enough. When trying
to read the next 4-byte word, StreamDevice will find at some point that
only one byte is left and stop reading. That byte (LF) will stay in the
input and you have to read it explicitly.
However, StreamDevice has one problem with GPIB but I don't know if it
affects you: The asynGpib implementation (at least for vxi11, the only
one I can test) does not allow to read the message in chunks. One has to
know in advance how long the message is and need to provide a
sufficiently large buffer. Therefore, StreamDevice starts with a 64 byte
buffer for GPIB devices. This is enough for most messages but not for
arrays. If the buffer is too small and the message cannot be read in
chunks, asyn should return a asynOverflow error. In that case,
StreamDevice increases the buffer size by doubling it. But this
transmission fails. One of the next transmissions may succeed as soon as
the buffer is sufficiently large.
This problem may be fixed in a future version of asyn or may even have
been fixed without me noticing it. One workaround would be to add a
variable to StreamDevice to specify the initial buffer size.
I have no idea how linux-gpib behaves. I can only test vxi11.
Gruss
Dirk
Hi all.
I've been trying to read waveforms from a Tektronix TDS8000 scope
using R3.14.11 + StreamDevice (2.4) + asyn (4.13) + linux-gpib using a
NI PCI GPIB interface on an x86 PC platform. I cannot seem to find the
right message termination to satisfy all of the various components.
The scope documentation says it does not use end-of-line terminator
characters, but rather sets EOI on the GPIB to signal end of messages.
Nevertheless, there are LF characters appended to each message, even
binary formatted waveform data. I'm trying to use RIB encoded
(Intel-endian binary) waveform data, 4-byte words.
I'm hoping to read 10 waveforms per second on a periodic scan of an
EPICS waveform record, (500 integer data points).
A simple bit of C code is able to read waveforms at about 80 waveforms
per second in a tight loop. In there, I have used the 'ibeos()'
function to cause my C code to not terminate reads on any character
delimiters, and this seems to have the desired effect. In the
gpib.conf file, I've include the setting:
set-reos = no
There is also the setting...
eos = ___
... which I've set to 0x0, commented out completely, and set to 0x0a,
none of which seems to change anything, and which is in accordance
with my expectation.
Most/some of the settings in gpib.conf seem to correspond to the
capabilities of the asynSetOption() function which I've also used in
the EPICS startup script. There, it does seem to make a difference. If
I set...
asynSetOption("GPIB0",10,0x0c,0)
...then asynTrace shows me that it does two things:
1. terminates its sequential reads on each binary zero byte
2. times out at the end of the read (and sees/reads the trailing LF)
I seem to be stuck between a rock and a hard place: either I use an
end-of-line terminator, which will eventually show up in the binary
data and prematurely terminate the read, or if I try to use the option
to not use end-of-message characters, it will wait until the read
times out, which seems to have the effect of invalidating the record,
and slowing things down a lot.
Any hints on how to proceed, tests to make or requests for further
info are welcomed.
Thanks.
Rod Nussbaumer
ISAC Controls, TRIUMF
Vancouver, Canada.
- Re: Reading scope waveforms with StreamDevice + asyn Dirk Zimoch
- Reading scope waveforms with StreamDevice + asyn Rod Nussbaumer
- Re: Reading scope waveforms with StreamDevice + asyn Rod Nussbaumer
RE: Strategies for working with fast-changing large arrays michael.abbott
<2012>
<2012>
