1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  <20102011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  <20102011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024
Subject: RE: aiRecord/devGpib problem
To: Andrew Wagner <[email protected]>, "[email protected]" <[email protected]>
Date: Fri, 6 Aug 2010 17:42:32 -0700
Hi Andrew,
> My first newbie question is why is SEVR: INVALID? The VAL is certainly what I want it
> to be.
The custom conversion routine (readMEAS) must reset the UDF field of the record (see devGpib.html in the section that describes GPIBREAD).  So you need to add this line:
>   pai->val = meas;
pai->udf = 0;
>   return 0;
Unfortunately, this won't solve your other problem with the sscanf not working properly.  BTW - you should check the return from sscanf.  Hopefully, someone else will immediately see the problem.  I have no gpib support like yours (with those embedded special characters).
Stephanie Allison
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf
> Of Andrew Wagner
> Sent: Friday, August 06, 2010 4:56 PM
> To: [email protected]
> Subject: aiRecord/devGpib problem
> Hey guys,
> I've got EPICS and ASYN running and have successfully followed one of Eric Norum's
> examples to talk to an Agilent 34970A MUX (over IP) and store the *IDN in a stringin
> record. I'm having problems however with reading a voltage measurement back reliably. I
> create device support with the devGpib template and the commands:
>   /* Param 0 -- Read identification string */
>   {&DSET_SI, GPIBREAD, IB_Q_HIGH, "*IDN?\n++read eoi",NULL, 0, 200, NULL, 0,
> 0, NULL, NULL, NULL},
>   /*Param 1 -- Read Measurement @ Channel */
>   {&DSET_AI, GPIBREAD, IB_Q_HIGH, "MEAS:VOLT:DC? (@203)\n++read eoi",
> NULL, 0, 200, readMEAS, 0, 0, NULL, NULL, NULL}
> I'm interfacing with the GPIB device via the PROLOGIX GPIB-Ethernet controller,
> (Google them, they're pretty neat) so you can ignore the ++read eoi part of the command
> string. The conversion method is:
> static int readMEAS( struct gpibDpvt *pdpvt, int P1, int P2, char **P3){
>   struct aiRecord *pai= (struct aiRecord *)pdpvt->precord;
>   double meas;
>   sscanf(pdpvt->msg,"%lg",&meas);
>   printf("meas %f\n",meas);
>   printf("Messge Length %d. Msg %s\n",pdpvt->msgInputLen,pdpvt->msg);
>   meas = (double)atof(pdpvt->msg);
>   pai->val = meas;
>   return 0;
> I set:
> asynSetTraceMask("L0",-1,0x9)
> asynSetTraceIOMask("L0",-1,0x2)
> in my st.cmd to do some debugging. When I force one of my voltage measurement records
> to process the first time I get:
> epics> dbtr AG34970:VOLT:Coil
> epics> 2010/08/06 15:39:30.848 172.25.100.85:1234 write 32
> MEAS:VOLT:DC? (@203)\n++read eoi\n
> 2010/08/06 15:39:31.011 172.25.100.85:1234 read 17
> +2.99865210E+00\n\003
> meas 2.998652
> Messge Length 15. Msg +2.99865210E+00
> meas 2.998652
> This is exactly what I expect since I'm supplying the MUX with 3 V at Channel 203. So far
> so good. When I print the record I get:
> epics> dbpr AG34970:VOLT:Coil
> ASG:                DESC: Coil Voltage  DISA: 0             DISP: 0
> DISV: 1             NAME: AG34970:VOLT:Coil                 RVAL: 0
> SEVR: INVALID       STAT: UDF           SVAL: 0             TPRO: 0
> VAL: 2.9986521
> to be. The problem occurs if I try to process the record more than once:
> epics> 2010/08/06 15:39:49.785 172.25.100.85:1234 write 32
> 2010/08/06 15:39:49.948 172.25.100.85:1234 read 17
> +2.99867890E+00\n\003
> meas 0.000000
> Messge Length 16. Msg +2.99867890E+00
> meas 0.000000
> Hmmm........an ASCII string indicating an appropriate value is read but converted into a
> double of 0.0000 . Note the message length has increased by 1. Its as if a hidden character is
> appended somewhere that screws up the sscanf and atof functions. Perhaps there is a more
> error proof way to convert ASCII to doubles? Has anyone seen this before? I have no such
> problem with my stringin record. I can send *IDN? and read the device name reliably.
> Thanks very much for your help. I will happily make any device support I get working
> public.
> Cheers,
> Andrew Wagner
aiRecord/devGpib problem Andrew Wagner
Prev: RE: aiRecord/devGpib problem Mark Rivers
Next: Re: BOY - XY Graph and macro substitution Ralph Lange
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  <20102011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024
Prev: Re: aiRecord/devGpib problem Andrew Wagner
Next: checkRelease on Windows ulrik.pedersen
