EXECUTIVE SUMMARY: Yes
DETAILS:
While looking at the severity set during the current round of record processing
it is nsev NOT sevr that should be checked.
A grep for sevr in src/rec for both 3.13 and 3.14 shows only two uses of sevr.
This one and in mbboDirect.c which has
if(pmbboDirect->nsev < INVALID_ALARM
&& pmbboDirect->sevr == INVALID_ALARM
Thus it is looking for the record going from INVALID to something else. This is OK.
I will fix this in both 3.13 and 3.14
Thanks again.
Marty Kraimer
Korhonen Timo wrote:
Hi Marty,
there was a little bit more to this story:
I wanted to use the IVOA field in calcout record to prevent the
NAN to be propagated to further records when the return value
from the calculation is invalid. This almost works - but the
first time the 'nan' gets through and corrupts other records
down the chain. It can even hang the ioc, and what was particularly
nasty that the nan was propagated down to a motor record and the
controller card (oms58) 'died', i.e., needed a sysreset to restart
it (after ctl-x, iocInit was hanging.)
After some more investigation (by the way, 3.14 on a Linux host is a
tremendous tool to debug this kind of things!) I could figure out that
the value really got only out once. The record decides if the outputs
are driven by looking at sevr field. However, sevr is the alarm status
from previous processing and gets updated only when recGblResetAlarms
is called (which in this case is done in the monitor routine.)
Now, my question is: is it OK to check nsev instead of sevr here?
I don't see a reason why not, but I may be missing something.
The code (around line 588 in calcoutRecord.c, in routine execOutput):
/* Check to see what to do if INVALID */
if (pcalc->sevr < INVALID_ALARM ) {
/* Output the value */
in the middle line, change pcalc->sevr to pcalc->nsev
This prevents the NAN being propagated (when the record is
configured properly.)
Timo
- calc/calcout and NAN Korhonen Timo
- Re: calc/calcout and NAN Marty Kraimer
- Re: calc/calcout and NAN Korhonen Timo
seq 1.9.5 restored to ftp site Peter Kurpis
<2002>
<2002>